OET 65 Supplement C EAB Part 22/24 SAR Review Item
1) Output power
a) Powers in SAR report must agree with EMC report and tune-up procedure
b) Conducted power in SAR report should be greater than or equal to what’s in EMC report, but not exceeding tune-up/tolerance
c) Scaling up or down 5% is allowed
d) Maximum output power (conducted) or SAR drift measured at same position in liquid before and after each SAR test?
2) Users Manual
a) Check for accessories and battery options – should test any accessories containing metal, and with all possible combinations; need justification if limited or “worst-case” combinations only tested
b) To comply with RF safety requirements use the specific belt clip. Other clips should be avoided and may not comply with RF safety requirements (for fair-trade do not exclude 3rd party acces.)
c) Check for RF Safety statements. Users must be clearly informed of the compliance requirements for device use, especially regarding body-worn configurations.
d) Check that FCC ID and SAR numbers in manual are correct
e) Do not allow any unsupported compliance claims
3) GENERAL REPORT INFORMATION
a)FCC ID listed (see 47 CFR §2.909)?
b)Statement of compliance with FCC RF exposure included (§2.1093)?
c)Mobile or portable transmitter device category identified?
d)Testing for Occupational/Controlled OR General Population/Uncontrolled limits?
e)Test device is production unit or identical prototype (47 CFR §2.908)?
4) Device operating configurations and test conditions
a)Brief description of the test device operating configurations included? For example:
i) Operating modes and operating frequency range(s)
ii) Maximum device rating for each operating mode and frequency range, both test sample and production units
iii) Antenna type and operating positions
iv) Applicable body-worn configurations
v) Battery options that could affect the SAR results
vi) Test positions for other accessories, e.g., earphones
b)Procedures to establish the test signals described (put phone on a call, e.g., base-station simulator vs internal test codes)? This may include a test equipment list or test codes
c)Multiple modes – CDMA, TDMA, GPRS, GSM, Bluetooth, etc?
d)Applicable source-based time-averaging duty factor and tested duty factor listed?
5) SAR Measurement system and site description
a)Brief description of the SAR scanning measurement system included?
b)Brief description of the test setup included: handset holders, phantom orientation, surroundings, absorber, noise floor, etc.
6)Electric field probe calibration
a)Description of the probe – including tip diameter, internal sensor offset from tip, etc
b)Description of the probe measurement errors included?
c)Description of probe calibration errors/uncertainties?
d)most recent calibration date and calibration certificate showing all factors used in report?
e)Check for consistency of probe factor and correspondent tissue parameters thru report
f)crest factor (peak-to-average voltage) parameters shown or needed, and/or addressed in calib.?